Here is the professional English translation of the calibration procedure for micro gear flowmeters after cleaning, incorporating key technical specifications and critical warnings:
🔧 Post-Cleaning Calibration Procedure
I. Pre-Calibration Preparation
Environmental Control
Maintain a dust-free, vibration-free environment with stable temperature (20–25°C/68–77°F) to prevent electromagnetic interference.
Equipment & Medium Preparation
Use calibration instruments with accuracy exceeding the flowmeter’s specifications (e.g., master flowmeters or volume tubes).
Select calibration fluid matching the actual fluid’s viscosity (e.g., high-viscosity lubricating oil); avoid gases or particulates.
Functional Pre-Check
Manually rotate gears to confirm smooth rotation with no jamming or noise.
Pressure-test seals using soap bubbles to detect leaks.
II. Core Calibration Steps
(1) Zero Calibration
Close valves to achieve static fluid conditions; stabilize for 5 minutes.
Adjust the zero-regulation valve or sensor position if readings deviate from zero (allowable drift: ≤±10 pulses).
Critical: Magnetic gear surface must face upward to ensure signal stability.
(2) Flow Range & Linearity Calibration
Segmented Flow Testing
Record readings at low (20%), medium (50%), and high (100%) flow rates against reference standards.
Gradually increase flow to avoid hydraulic shock on gears.
Deviation Correction
For persistent errors >±0.5%:
Measure gear-to-wall clearance with a micrometer (multipoint average).
Loosen mounting screws, realign gears for frictionless rotation, and retighten.
Hysteresis Verification
Compare forward/reverse flow readings; hysteresis error should be <0.2%.
(3) Accuracy Validation Methods
Method Key Steps
Volumetric Method Collect fixed fluid volume; compare with flowmeter’s totalized value
Master Meter Method Connect a reference flowmeter in series; synchronize readings
Acceptance Criteria: Error ≤±0.5% at all test points with repeatable results.
⚠️ III. Post-Cleaning Critical Checks
Residual Solvent Removal
Eliminate cleaning agents (e.g., isopropanol) completely to prevent viscosity changes.
Gear Reassembly Validation
Rotate gears manually 3–5 times to verify proper meshing.
Reinstall gears if magnetic surfaces are misaligned.
Seal Integrity Retest
Replace all disassembled seals; tighten cover screws diagonally to prevent distortion.
IV. Post-Calibration Maintenance
Calibration Cycle: Annually for general use; every 3–6 months in high-contamination/high-use environments.
Quarterly Verification: Compare with portable master meters; recalibrate if deviation >0.5%.
Failure Analysis: Replace gear sets if linearity errors persist (indicates eccentric wear).
Run the flowmeter for 24 hours post-calibration to confirm operational stability.
Key Terminology Translations:
零点校准 → Zero calibration
线性度 → Linearity
回程误差 → Hysteresis error
满管无气穴 → Bubble-free pipe filling
密封圈 → Seals/O-rings